Question 1 Multiple Choice
A clinical lab uses a direct ISE to measure K⁺ in undiluted blood plasma. The measured K⁺ activity is 3.8 mM, but the total K⁺ concentration determined by atomic absorption is 4.2 mM. Why does this discrepancy exist?
AThe ISE is malfunctioning because blood plasma contains too many interfering ions that overwhelm the K⁺ selectivity
BIn high ionic strength solutions like plasma, ion-ion interactions reduce the effective concentration the electrode senses — the ISE measures activity, not total concentration, and these diverge at high ionic strength
CThe ISE has been incorrectly calibrated with low-ionic-strength aqueous standards instead of plasma-matched standards
DK⁺ in blood plasma is partially bound to albumin and not freely ionized, which the ISE cannot detect
Ion-selective electrodes respond to ion activity, not total concentration. Activity = γ × c, where γ is the activity coefficient and c is the concentration. In dilute solutions γ ≈ 1 and activity ≈ concentration. In high ionic strength matrices like blood plasma, ion-ion interactions cause γ < 1, so activity is systematically lower than concentration. The ISE faithfully reports activity (3.8 mM) while atomic absorption reports total concentration (4.2 mM); neither is wrong — they measure different quantities. Understanding which one your method gives is essential for clinical interpretation.
Question 2 Multiple Choice
The glass pH electrode measures pH without consuming any reagent or altering the sample. This non-destructive character arises because:
AThe glass membrane catalyzes a reversible acid-base reaction that regenerates itself, consuming no net reagent
BNo electrical current flows through the measurement circuit — a high-impedance voltmeter detects the potential difference across the glass membrane without driving any electrochemical reaction in the sample
CThe internal reference buffer solution neutralizes any pH changes caused by the measurement, restoring the sample
DH⁺ ions are temporarily absorbed into the glass and then released back to the solution after measurement
Potentiometric measurement is a zero-current technique. A high-impedance voltmeter draws essentially no current, so no electrochemical reactions are driven in the sample and no analyte is consumed or altered. The glass membrane develops a potential difference due to H⁺ activity differences on its two faces, but this is an equilibrium property that does not require net ion transfer. This is why ISE measurements can be made in very small, precious, or reactive samples — the electrode reads the signal without changing the system.
Question 3 True / False
Ion-selective electrodes measure the activity of the target ion, which equals its molar concentration in most aqueous solutions at standard conditions.
TTrue
FFalse
Answer: False
Activity equals concentration only in the limit of infinite dilution, where activity coefficient γ → 1. At any real ionic strength, activity = γ × c, and γ < 1 due to electrostatic interactions among ions. For clinical samples (blood plasma, urine) or environmental samples (seawater, concentrated soil extracts), ionic strength is high enough that activity and concentration differ meaningfully. Failing to account for this is a common source of error when interpreting ISE results alongside concentration-based reference methods.
Question 4 True / False
The Nernst equation predicts that a tenfold change in H⁺ activity (one pH unit) produces a change of approximately 59.2 mV in the glass pH electrode potential at 25°C.
TTrue
FFalse
Answer: True
The Nernst equation for the glass electrode is E = E° + (RT/F) × ln(a_H⁺). At 25°C, RT/F = 25.7 mV, and multiplying by ln(10) ≈ 2.303 gives the Nernstian slope: 59.2 mV per decade change in activity (i.e., per pH unit). A real glass electrode's slope deviates slightly from this theoretical value due to membrane imperfections; calibration with two buffers determines the actual slope and intercept for that specific electrode. The theoretical 59.2 mV/pH is the reference expectation.
Question 5 Short Answer
What is the difference between ion activity and ion concentration, and why does this distinction matter for interpreting ISE measurements in biological or environmental samples?
Think about your answer, then reveal below.
Model answer: Ion concentration is the total amount of an ion per unit volume (mol/L). Ion activity is the thermodynamically effective concentration — the concentration corrected for the non-ideal behavior of ions in solution: activity = γ × concentration, where γ is the activity coefficient. In dilute solutions γ ≈ 1 and the two are essentially equal. In high ionic strength matrices (blood plasma, seawater, concentrated buffers), ions interact electrostatically with each other, reducing their effective activity so that γ < 1. ISEs respond to activity, not concentration. This matters because a blood K⁺ ISE reading of 3.8 mM (activity) corresponds to a somewhat higher total concentration; using the activity value directly as a concentration when interpreting against concentration-based normal ranges can introduce clinically significant errors. Clinical analyzers handle this either by diluting samples (making γ ≈ 1) or by calibrating against activity-matched plasma standards.
The activity vs. concentration distinction also affects method comparison: atomic absorption and flame photometry measure total concentration; ISEs measure activity. When results from these methods are compared in the same sample, the difference reflects ionic strength effects, not measurement error.